This is an older Mitu CD6-CSX that has seen only light use. Have a job here setting shims in mechanical injectors so I wanted a quick and reliable way to measure them from stacks that I have laying around. Quick yes. Reliable, not so much. They weren't measuring to known stock thicknesses with the Mitu so I checked against my mic and vernier. Both of those read correctly, but the Mitu is about 2 thou or .05 mm low quite consistently. Replaced battery; checked for jaw parallel, multiple reset of zero -- all good. Pulled out my 1 inch and 2 inch standards. Same thing, reading .0025 low at 1 inch and .0020 at 2 inch. I know the last digit is fragile.
I assume these things are constructed with glass scales so if they work, shouldn't they be accurate? Is there a reset or calibration procedure?
I have used it a lot for the sheer convenience, but now that I can't trust it, I guess it's back to the old Mitu vernier for a while.
I assume these things are constructed with glass scales so if they work, shouldn't they be accurate? Is there a reset or calibration procedure?
I have used it a lot for the sheer convenience, but now that I can't trust it, I guess it's back to the old Mitu vernier for a while.